QDR-4000 Failure.

(11)
(5)

Alejandro Oyola

QDR-4000 Failure.

August 18, 2010 10:27

When I ask for a QA test the software shows the message "Calib: time out".
I checked the A/D to T&C cable,The DC voltages All OK.
With the survey program I checked that the x ray system,A/D board and detector functions.
Please any ideas to fix it.

Log inor Register

to rate and post a comment

(1)

William Deery

Re: QDR-4000 Failure.

August 26, 2010 10:14

Check the filter wheel pick off adjustment. Sounds like the system can't sync with the filter wheel. The filter wheel on the tank should be spinning and you should be able to scope the signal coming back from the pick-off sensor. The signal is fed to the XRC. That could be your probelem, a bad XRC. Good luck.

Log inor Register

to rate and post a comment

(11)
(5)

Alejandro Oyola

Re: QDR-4000 Failure.

August 27, 2010 01:54

Thank you for your assistance,I checked the pick off and sync with the survey program and all values appear ok as shows the service manual.
I continue working.