qxi scanning hardware reset failed

January 09, 2017 08:49

Hello May be somebody can help me, I have a light speed qxi with this error: Scanning hardware reset failed. I made ping with etc stc and obc and answer is unknown Host.

Log inor Register

to rate and post a comment

(62)
(11)

zekeng brice

re: qxi scanning hardware reset failed

January 09, 2017 01:21

hello juan you need to specify the error messages and let know the type of console your machine has octane [green] or octane 2 blue as you are presenting the thing you have communication issue

Log inor Register

to rate and post a comment

(7)

(4)

Juan Jose Ventura

re: re: qxi scanning hardware reset failed

January 10, 2017 10:12

hello Brice thank you for your answer
The system have green octane the SBC is OS computer.
The problem is the communication between console to scanner. I checked converter UTP to coaxial is ok.
error report
Host: ct01_sbc
Function: No system function reported
Scan Type: Not scanning Scan:0/0/0
Scan Seq Id: 0 The STC controller has failed to reset



Host:......
the Same


Scan Seq Id: 0 The reset failure of the OBC may be due to the reset failure of the STC


Host:.......
The same


Scan Seq Id: 0 he ETC controller has failed to reset


Log inor Register

to rate and post a comment

(23)
(4)
(9)

Deepak Roy Sharma

re: qxi scanning hardware reset failed

January 10, 2017 10:16

Check for Media converter inside Console if its working fine

Log inor Register

to rate and post a comment

(7)

(4)

Juan Jose Ventura

re: re: qxi scanning hardware reset failed

January 11, 2017 10:20

De media converter working good thank you

Log inor Register

to rate and post a comment

(11)
(3)

AKMAL TURAEV

re: qxi scanning hardware reset failed

January 10, 2017 10:17

Hi Juan, did you check the connection? Mostly you have got an communication error. check the connection of the console and from left side of the gantry check the STC and OBC boards lights are they ok or not.

Log inor Register

to rate and post a comment

(7)

(4)

Juan Jose Ventura

re: re: qxi scanning hardware reset failed

January 11, 2017 10:21

the Gantry have three ,modules ETC STC OBC When I do the ping the system return unknown host in the three modules. These modules are not independent?
at leasts ETC STC?