Hello everyone!
I got a problem with my Radiometer ABL800. The Hemolyzer test in Service menu shows these results:
1. Thermostat temperature: Passed; 37.2 degrees
2. Thermostat heat: Passed 12.5%
3. Tuning voltage: Failed 67.0 Vpeak
4. Tuning Frequency: Passed 30.284khz
5. Impedance shape: Passed
6. Cavitation absorbance: Failed Max 4mAbs
7. Cavitation Drive Current Failed 100%

I believe that "pass" tests are mainly related to the logicboard/controller and the "fail" tests are related to the mechanics of the hemolyzer.

What do you think? Is it possible to fix it?

Thank you :)

Log inor Register

to rate and post a comment

(172)
(40)
(3)

Dmitry Dederer

re: Radiometer ABL 800 hemolyzer failure

November 16, 2014 11:19

Hello! Do you have service manuals for this system? You need to check caviation. What kind of gas do you use?

Log inor Register

to rate and post a comment

(5)
(2)

Ciprian Lupu

re: re: Radiometer ABL 800 hemolyzer failure

November 20, 2014 06:30

Hello,I do have a service manual for this system but it does not contain any info about the hemolyzer. Maybe Radiometer keeps this kind of info only for themselves :-) Unfortunately I cannot understand your question . Are you refering to the calibration gas? Have a nice day

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: Radiometer ABL 800 hemolyzer failure

July 22, 2015 11:33

Dmitry
do you have the service manual for this unit? i am trying to help someone and the manual would be helpful
thank you

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)