Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Regius 110 Remove Cassete

Tags: PACS / DR / CR / Digital, KONICA MINOLTA

(3)

Evgeny Sazonov

Regius 110 Remove Cassete

September 07, 2018 10:39

Hello! When you turn on the device, writes Remove Cassette. What kind of sensor? Does anyone have a service manual. Thank you. Sorry for my English. My Mail e_saz@mail.ru

Log inor Register

to rate and post a comment

(84)
(16)

fortronik

re: Regius 110 Remove Cassete

September 08, 2018 03:06

Checking the the Cassette insertion transaptors can be dirty or damaged

1 person liked this
Log inor Register

to rate and post a comment

(3)

Evgeny Sazonov

re: re: Regius 110 Remove Cassete

September 09, 2018 11:16

Thank you. I cleaned the photo sensor. Everything worked.

Log inor Register

to rate and post a comment

(3)

Evgeny Sazonov

re: re: Regius 110 Remove Cassete

September 09, 2018 11:16

Thank you. I cleaned the photo sensor. Everything worked.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)