Nikolas
Regius 210 Error13000
June 12, 2024 10:24
Hi, i have the error 13000 after starting and selftest. Anybody know what's the reason?
may be share some service manual, thanks
to rate and post a comment
fortronik
re: Regius 210 Error13000
June 13, 2024 11:16
error 13000 -H-sync timeout
No H-sync signal could be received during reading.
Restart the unit. If the error reoccurs, check the following:
1) Check the wiring for the SCB3/ADB2 board.
2) Replace the optical unit and check operation.
to rate and post a comment
Nikolas
re: Regius 210 Error13000
June 13, 2024 11:16
I have changed oprical module, its still the same problem.
to rate and post a comment
Nikolas
re: re: Regius 210 Error13000
June 17, 2024 10:39
Remedy: exchanged wires between OU and adb2.