RF Problem

Tags: MRI

Nuhad Salman

RF Problem

April 13, 2017 10:43

I have 0.2T Signa EXCITE, Iam having a problem with the RF, during scanning the RF goes and the error messages shows (RF Recovery in progress) we did fine tunning for the coil and we did Gain calibration test, but when the RF goes the fans of the RF Amplifier stop, after one minute the RF Amplifier goes back to ready state and we can continue the scan, but recently the RF issue if tooooo frequent and some system cant work anymore. Any one can give me the suspected fault? Thank you

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: RF Problem

April 14, 2017 06:03

What is the error logs? Did you make nanual prescan?

Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: RF Problem

April 16, 2017 08:44

Either the input power to the RF amplifier is not secure or, the power supplies in the RF amplifier are going bad. We can repair the amplifier for you. bruce@medsystech.com

Log inor Register

to rate and post a comment

(21)

INOWER

re: RF Problem

April 25, 2017 08:19

It seems like it is related to RF amp Share any error logs if possible. Can support you to rectify the problem.

inowerelectricals@gmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We repair MRI Coils, RF amplifiers, Gradient Amplifiers and Injectors.

MIT labs, experts in Multi-Vendor component level repair of: MRI Coils, RF amplifiers, Gradient Amplifiers Contrast Media Injectors. System repairs, sub-assembly repairs, component level repairs, refurbish/calibrate. info@mitlabsusa.com/+1 (305) 470-8013

stats
advertisement

Upcoming Auction Events

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (218 lots)