I am looking for documentation on a frequency conversion kit for a GE RFX/SFX x-ray table.
I have the catalog number but GE was unable to produce usable information reguarding the conversion from 60Hz to 50Hz input power for this table.
Any information received will be appreciated
to rate and post a comment
Joseph Jenkins
Re: RFX/SFX FREQUENCY CONVERSION
May 16, 2002 12:23
Good day,
Concerning your request for info on 60hz to 50hz conversion.
1) If you get any info i very much would like a copy>>>iiltd@hotmail.com
2) I in stalled a RFX II into a 50Hz country and I found that
a) There is 1 jumper in the table cab. on the table controller card(bottom,straight back ,center)that is to be placed into the 50HZ position.
b) the 26V supplies may not (there are 4)put out the correct voltage>>1 out of that i had did not work( the xfor are resinant at 60hz)so i n stall a hi current(circuit is fused at 6Amp) 26vdc supply(out of a ultrasound system) at the output of that supply and everything else has worked fine.
Let me know how it works out and send me any info you can find, if you can find any>>>only GE england will have anything.
Jay Jenkins