Recently, I found a problem with Toshiba General X-Ray KXO-50F high speed starter. (Model: RS-200)
When Shooting High KV, intermittent no X-ray.
I found high speed starter show error two dot (2.) at seven segment display. I also found a symptom that the segment segment display of high speed starter change from 0 to 1 and jump back to 0 and the cooling fan of high speed starter continue run and stop.
I tried to change the CPU board but still happen.
Anyone have the idea to solve the problems?
At the bottom, the printed circuit, often breaking track piloting power transistor. You have to do continuity test, because the disruption is not visible.
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
yikwang
RS-200 High speed starter problem
September 25, 2014 06:51
Recently, I found a problem with Toshiba General X-Ray KXO-50F high speed starter. (Model: RS-200)
When Shooting High KV, intermittent no X-ray.
I found high speed starter show error two dot (2.) at seven segment display. I also found a symptom that the segment segment display of high speed starter change from 0 to 1 and jump back to 0 and the cooling fan of high speed starter continue run and stop.
I tried to change the CPU board but still happen.
Anyone have the idea to solve the problems?
to rate and post a comment
Lenny Place
re: RS-200 High speed starter problem
September 27, 2014 08:14
If I remember correctly the 2. means the stator start signal is present at power on.
check for shorted FET's on the amp modules
to rate and post a comment
DIOGENES MARCONDES
re: RS-200 High speed starter problem
April 11, 2015 09:03
At the bottom, the printed circuit, often breaking track piloting power transistor. You have to do continuity test, because the disruption is not visible.