Over 500 Total Lots Up For Auction at One Location - NJ 06/15

rt3200 II white screen

(3)

Michael Melton

rt3200 II white screen

August 15, 2022 10:18

Hello!
When I turn on the rt3200, i get a white screen. Behind the white screen, i can barely make out words in the top left corner. Juat appears as if the while screen has a white overlay.

I have replaced the battery on the MST board and have done a reset, but no success. And help is appreciated!

Log inor Register

to rate and post a comment

(32)
(4)

Tiorico_86

re: rt3200 II white screen

August 16, 2022 10:54

You must to check the power supplies, I had the same problem and no had the 5vcd. I changed and everything was ok.

Log inor Register

to rate and post a comment

(3)

Michael Melton

re: re: rt3200 II white screen

August 17, 2022 09:51

Okay thank you!! Did you do it yourself? Or send it in? Not sure which the 5vcd is

Thanks you again!