Chiller working fine 26C to 30 C, IDC says: Power detector off 1mn and then tries to power it on many times. Firnware corrupted. I did pull the FW from the IDC, it didnt work, I updated the detector files, it didnt work. I checked IDC batery and BIOS config, both OK. I run a restore to IDC from backup I had, it didnt work. Coolant level is OK, no licking found. Help
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Douglas Varela
S83 Detector environment error
October 24, 2022 10:25
Chiller working fine 26C to 30 C, IDC says: Power detector off 1mn and then tries to power it on many times. Firnware corrupted. I did pull the FW from the IDC, it didnt work, I updated the detector files, it didnt work. I checked IDC batery and BIOS config, both OK. I run a restore to IDC from backup I had, it didnt work. Coolant level is OK, no licking found. Help
to rate and post a comment
Carlos Duran
re: S83 Detector environment error
October 25, 2022 10:30
Hello Douglas,
Check chiller with flow meter if is into specs. It seems that aged chiller doesn't work as supposed to be.
There is a Job Card in the service manual with the indications