S83 Detector environment error

Tags: Mammo, GE

(22)

Douglas Varela

S83 Detector environment error

October 24, 2022 10:25

Chiller working fine 26C to 30 C, IDC says: Power detector off 1mn and then tries to power it on many times. Firnware corrupted. I did pull the FW from the IDC, it didnt work, I updated the detector files, it didnt work. I checked IDC batery and BIOS config, both OK. I run a restore to IDC from backup I had, it didnt work. Coolant level is OK, no licking found. Help

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: S83 Detector environment error

October 25, 2022 10:30

Hello Douglas,

Check chiller with flow meter if is into specs. It seems that aged chiller doesn't work as supposed to be.

There is a Job Card in the service manual with the indications