Manuel Esquer

Saved RAM Data lost on CTe

November 09, 2015 10:26

I have a problem with GE CTe (O2), XG ERROR 90-701H, Application error: Saved RAM checksum pb
I'll tried run Checksum on Generator Test but not solved
Then I'll run Saved RAM data Download but send Error: File open failure.
After that I installed Software in a new HDD, but still same problem
and finally I replace NVRAM and problem chance 90-705H, but Error: File open failure, trying to run Saved RAM Download still appears too
Any comment about it? Thank you

2 people liked this
Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: Saved RAM Data lost on CTe

November 10, 2015 08:36

Error code: 90-0701H
Message/explanation: corruption in the NV Ram verification (After power up)
Potential cause:
1. Database problem
2. External cause (Spikes)
3. kV control
Recommended action/ Troubleshooting guide:

1 person liked this