When try to run a QA and start to check secondary block position, moves to center of table, when it start to move again to read block, stops and send SCANNER RESET ERROR.
Any idea?
Who knows how interpret errors?
Thank you in advance
That sounds like it could be a broken wire in the harness or a poor connection at the lower arm where the cable bulkhead connectors are. If you have the error and code I may have a manual that might have more info on the error. It's been quite a while scene I have worked on this model scanner.
Dear Villa please be more spesific. First of all NT doesnt go to center of the table in the begining of QA test. Goes above the standard black block. Opens lazer on, and expects to set the black box to the correct position.
Scanner reset error is a general error which means everything and nothing. You have to see all the context.
Generaly means something crucial is wrong and I have to stop communication and start again...
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
villa
SCANNER RESET ON DPX-NT
August 17, 2020 11:11
When try to run a QA and start to check secondary block position, moves to center of table, when it start to move again to read block, stops and send SCANNER RESET ERROR.
Any idea?
Who knows how interpret errors?
Thank you in advance
to rate and post a comment
ALLEN GODFREY
re: SCANNER RESET ON DPX-NT
August 19, 2020 10:35
That sounds like it could be a broken wire in the harness or a poor connection at the lower arm where the cable bulkhead connectors are. If you have the error and code I may have a manual that might have more info on the error. It's been quite a while scene I have worked on this model scanner.
to rate and post a comment
ALFATECH
re: SCANNER RESET ON DPX-NT
August 24, 2020 10:43
Dear Villa please be more spesific. First of all NT doesnt go to center of the table in the begining of QA test. Goes above the standard black block. Opens lazer on, and expects to set the black box to the correct position.
Scanner reset error is a general error which means everything and nothing. You have to see all the context.
Generaly means something crucial is wrong and I have to stop communication and start again...
to rate and post a comment