I can do the CR scan properly but during normal scan (rotation mode) I got at the end : DATA AQ ERROR ! then system go to Ready mode and no image is displayed.
Any suggestion is strongly appreciated.
Dear Ralph, Im from Mongolia, our hospital has same problem as Alessandro. How did you fix it in 2004. Sorry for asking after 10 years. Hope that you can help me!
most probably it is due to your rotational ckty.
it could be your encoder, or a rusty switch on drum. you can diagnose the defective switch it by rotating the gantry with local control box.
GCI offers complete MRI & CT sales & service with ROI analysis, in-house financing, contractor drawings, accreditation, training, marketing, and OEM-level support beyond the sale.
Alessandro Lazzari
SCT 4800 TF trouble : data AQ error
November 18, 2004 10:03
I can do the CR scan properly but during normal scan (rotation mode) I got at the end : DATA AQ ERROR ! then system go to Ready mode and no image is displayed.
Any suggestion is strongly appreciated.
to rate and post a comment
Ralph Childs
Re: SCT 4800 TF trouble : data AQ error
November 18, 2004 12:22
Alessandro,
Your problem is related to rotational das pulses from the gantry. You need to check the encoder and also the input signal to the das trigger.
You can reach me at scanner1@eim.ae
to rate and post a comment
Munkhtuul Togooch
re: Re: SCT 4800 TF trouble : data AQ error
March 11, 2014 08:05
Dear Ralph, Im from Mongolia, our hospital has same problem as Alessandro. How did you fix it in 2004. Sorry for asking after 10 years. Hope that you can help me!
to rate and post a comment
Iftikhar Khan
Re: SCT 4800 TF trouble : data AQ error
November 19, 2004 12:09
most probably it is due to your rotational ckty.
it could be your encoder, or a rusty switch on drum. you can diagnose the defective switch it by rotating the gantry with local control box.
u can contact me on iftekhar786@hotmail.com
to rate and post a comment
Carlos Duran
Re: SCT 4800 TF trouble : data AQ error
November 21, 2004 11:48
hello:
if the CR image is ok, the QV boards are ok, and detector too.
i think that you must run the software for check data acquisition under rotation, to see how send info into movement.
the das cable would be broken (some lines), or encoder could be bad too.
regards,
carlos
to rate and post a comment
Edwin Rosado
Re: SCT 4800 TF trouble : data AQ error
November 23, 2004 06:38
hi:
if you send me your email, i send you some instruction about how test troubleshoting this problem. your situation is in the encoder area.
to rate and post a comment