Hello
Iam facing a problem with a SECHRIST IV100B INFANT VENTILATOR. The device doesnt cycle properly, it does not supply the inspiratory presure, the exhalation presure outputs dont work normaly. Can anyone help with this issue?.
Thanks beforehand
Ruben
From skilled GE depot repairs to patient-ready accessories, Southeastern Biomedical specializes in GE to keep your facility safe, efficient, and cost-effective. 888-310-7322.
Ruben Cadima
SECHRIST IV100B PROBLEM
June 13, 2012 11:43
Hello
Iam facing a problem with a SECHRIST IV100B INFANT VENTILATOR. The device doesnt cycle properly, it does not supply the inspiratory presure, the exhalation presure outputs dont work normaly. Can anyone help with this issue?.
Thanks beforehand
Ruben
to rate and post a comment
Enrique Rozzi
re: SECHRIST IV100B PROBLEM
June 14, 2012 06:03
disarm de manifold and change the valves called duckbill. That´s all.
to rate and post a comment
Chris Baklik
re: SECHRIST IV100B PROBLEM
June 15, 2012 09:45
Ruben
I have the service manual e-mail me crisb1955@aol.com
to rate and post a comment