Sedecal global showing E22

(63)
(7)
(6)

Punit Singh

Sedecal global showing E22

September 01, 2016 10:40

I have old sedecal global which is showing E22 error on switch on. "Tube ovetheat" Thereafter no other key is operational. Ie console freeze

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Sedecal global showing E22

September 02, 2016 03:38

I thought E22 was wrong tube selected tube 2?

2 people liked this
Log inor Register

to rate and post a comment

(63)
(7)
(6)

Punit Singh

re: re: Sedecal global showing E22

September 03, 2016 05:01

Yes its showing tube 2 but unable to select tube 1. How we can reset this as this comes on as soon as we start

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Sedecal global showing E22

September 06, 2016 04:06

you could of lost the battery backup in the generator on the controller board which means you need to configure the system.
if you need a service manual please send me an email address

Log inor Register

to rate and post a comment

(44)
(2)

villa

re: re: re: re: Sedecal global showing E22

September 06, 2016 06:26

I think he is able to recover all info cal by it self, after change battery

Log inor Register

to rate and post a comment

(63)
(7)
(6)

Punit Singh

re: re: re: re: Sedecal global showing E22

September 07, 2016 12:17

punit.globalmedical@gmail.com yes manual would be great help

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: Sedecal global showing E22

September 07, 2016 03:16

i emailed it to you

1 person liked this
Log inor Register

to rate and post a comment

(4)

Lengyel Akos

re: re: re: re: re: re: Sedecal global showing E22

September 24, 2017 10:13

Please send it to me as well

Log inor Register

to rate and post a comment

(4)

Lengyel Akos

re: re: re: re: re: re: Sedecal global showing E22

September 24, 2017 10:13

Please send it to me as well.
drtapfesz@hotmail.com

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: re: re: Sedecal global showing E22

September 25, 2017 09:09

ok

Log inor Register

to rate and post a comment

Austin Cole

re: Sedecal global showing E22

April 04, 2019 09:46

What did this end up being? I have same issue

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: Sedecal global showing E22

April 05, 2019 09:13

DESCRIPTION : Wrong Tube-2 selection.
ERROR TYPE : Informative. Does not allow operation until the error is solved or another Tube is selected.
APPLICABLE TO : All Generators.
APPEARS WHEN : At any moment mainly after initialization or when changing Tube selection.
INFORMATION / SYMPTOM : Wrong Tube 2 selection.

POSSIBLE CAUSES
The –HT INTLK is not at high logic level 5V (Tube 2 selected).
Defective HV Switch inside the HV Transformer (two Tubes option).
The --HT INTLK is grounding.

ACTIONS
Note: Before performing any action, check the Workstation configuration for each Tube.
Only for one (1) Tube option:
1. Check the –HT INTLK in 6J3-13 (Generator Cabinet), it should be 5 VDC with Tube 2 selected.
2. If it is not 5 VDC, check that 6J3-13 is not grounding.
3. If 6J3-13 = 5 VDC, check J3--13 on the ATP Console CPU Board.
4. If not 5 VDC in J3-13, disconnect J3--13 on the ATP Console CPU Board and check for
short--circuits between J3--13 and ground or others signals.
5. If voltage is 5 VDC and “E22” appears, replace the ATP Console CPU Board.
Only for two (2) Tubes option:
1. Check J3-13 on the ATP Console CPU Board. If it is 5 VDC, replace the ATP Console CPU
Board. If it is 0 VDC, proceed to the second step.
2. Turn off the equipment, disconnect J1 from the HV Transformer and check for short--circuits
between J1--J and ground or others signals.
3. If previous steps are correct, replace the HV Transformer.

Log inor Register

to rate and post a comment