you could of lost the battery backup in the generator on the controller board which means you need to configure the system.
if you need a service manual please send me an email address
DESCRIPTION : Wrong Tube-2 selection.
ERROR TYPE : Informative. Does not allow operation until the error is solved or another Tube is selected.
APPLICABLE TO : All Generators.
APPEARS WHEN : At any moment mainly after initialization or when changing Tube selection.
INFORMATION / SYMPTOM : Wrong Tube 2 selection.
POSSIBLE CAUSES
The âHT INTLK is not at high logic level 5V (Tube 2 selected).
Defective HV Switch inside the HV Transformer (two Tubes option).
The --HT INTLK is grounding.
ACTIONS
Note: Before performing any action, check the Workstation configuration for each Tube.
Only for one (1) Tube option:
1. Check the âHT INTLK in 6J3-13 (Generator Cabinet), it should be 5 VDC with Tube 2 selected.
2. If it is not 5 VDC, check that 6J3-13 is not grounding.
3. If 6J3-13 = 5 VDC, check J3--13 on the ATP Console CPU Board.
4. If not 5 VDC in J3-13, disconnect J3--13 on the ATP Console CPU Board and check for
short--circuits between J3--13 and ground or others signals.
5. If voltage is 5 VDC and âE22â appears, replace the ATP Console CPU Board.
Only for two (2) Tubes option:
1. Check J3-13 on the ATP Console CPU Board. If it is 5 VDC, replace the ATP Console CPU
Board. If it is 0 VDC, proceed to the second step.
2. Turn off the equipment, disconnect J1 from the HV Transformer and check for short--circuits
between J1--J and ground or others signals.
3. If previous steps are correct, replace the HV Transformer.
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
Punit Singh
Sedecal global showing E22
September 01, 2016 10:40
I have old sedecal global which is showing E22 error on switch on. "Tube ovetheat" Thereafter no other key is operational. Ie console freeze
to rate and post a comment
Lenny Place
re: Sedecal global showing E22
September 02, 2016 03:38
I thought E22 was wrong tube selected tube 2?
to rate and post a comment
Punit Singh
re: re: Sedecal global showing E22
September 03, 2016 05:01
Yes its showing tube 2 but unable to select tube 1. How we can reset this as this comes on as soon as we start
to rate and post a comment
Lenny Place
re: re: re: Sedecal global showing E22
September 06, 2016 04:06
you could of lost the battery backup in the generator on the controller board which means you need to configure the system.
if you need a service manual please send me an email address
to rate and post a comment
villa
re: re: re: re: Sedecal global showing E22
September 06, 2016 06:26
I think he is able to recover all info cal by it self, after change battery
to rate and post a comment
Punit Singh
re: re: re: re: Sedecal global showing E22
September 07, 2016 12:17
punit.globalmedical@gmail.com yes manual would be great help
to rate and post a comment
Lenny Place
re: re: re: re: re: Sedecal global showing E22
September 07, 2016 03:16
i emailed it to you
to rate and post a comment
Lengyel Akos
re: re: re: re: re: re: Sedecal global showing E22
September 24, 2017 10:13
Please send it to me as well
to rate and post a comment
Lengyel Akos
re: re: re: re: re: re: Sedecal global showing E22
September 24, 2017 10:13
Please send it to me as well.
drtapfesz@hotmail.com
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: Sedecal global showing E22
September 25, 2017 09:09
ok
to rate and post a comment
Austin Cole
re: Sedecal global showing E22
April 04, 2019 09:46
What did this end up being? I have same issue
to rate and post a comment
Lenny Place
re: re: Sedecal global showing E22
April 05, 2019 09:13
DESCRIPTION : Wrong Tube-2 selection.
ERROR TYPE : Informative. Does not allow operation until the error is solved or another Tube is selected.
APPLICABLE TO : All Generators.
APPEARS WHEN : At any moment mainly after initialization or when changing Tube selection.
INFORMATION / SYMPTOM : Wrong Tube 2 selection.
POSSIBLE CAUSES
The âHT INTLK is not at high logic level 5V (Tube 2 selected).
Defective HV Switch inside the HV Transformer (two Tubes option).
The --HT INTLK is grounding.
ACTIONS
Note: Before performing any action, check the Workstation configuration for each Tube.
Only for one (1) Tube option:
1. Check the âHT INTLK in 6J3-13 (Generator Cabinet), it should be 5 VDC with Tube 2 selected.
2. If it is not 5 VDC, check that 6J3-13 is not grounding.
3. If 6J3-13 = 5 VDC, check J3--13 on the ATP Console CPU Board.
4. If not 5 VDC in J3-13, disconnect J3--13 on the ATP Console CPU Board and check for
short--circuits between J3--13 and ground or others signals.
5. If voltage is 5 VDC and âE22â appears, replace the ATP Console CPU Board.
Only for two (2) Tubes option:
1. Check J3-13 on the ATP Console CPU Board. If it is 5 VDC, replace the ATP Console CPU
Board. If it is 0 VDC, proceed to the second step.
2. Turn off the equipment, disconnect J1 from the HV Transformer and check for short--circuits
between J1--J and ground or others signals.
3. If previous steps are correct, replace the HV Transformer.
to rate and post a comment