I NEED HELP, IN SELECT SP, EX.
70KV X 70MA 2SECONDS, VALUE OF FEEDBACK IN ANODE CUURENT IS OUT
PLEASE HELP ME
PLEASE I NEED EMAIL OF Wojciech Selwesiuk
to rate and post a comment
Leungbun Liu
Re: Select SP problem
October 08, 2009 10:13
What error you get in SCU ? Should starting with 3XXX.
I assume you are using the diagnostic mode to make a exporure of 140KV 70ma, and u are working on the RMS Card diag switch and also the switch on the Anode side to disable the cathode side. Right ?
Give me more details and maybe I can help ?
to rate and post a comment
Wojciech Selwesiuk
Re: Select SP problem
October 09, 2009 05:23
Hi,
1. Hmmm, as I know, SeleCT doesn't work with 70kV. Default voltage in scu_diag application is 80kV 100mA 2100ms. You can change them but available are: 80,90,100,110,120,130,140.
2. Have you got any anode current values or they are crazy?
wselwesiuk@synektik.com.pl
to rate and post a comment