the error 5 means "calculated exposure time less than a minimum" the exposure is too short. under what conditions this error occur? the AEC sensor is on automatic or it is on one of the positions? bye
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
juan pablo amado
selenia aec error
October 02, 2014 04:47
intermitent but repetitive error 5
log:
What is the Xtm value?
Xcount M F kV mAs S f R A C D AEC Xtm mA1 mA2 FIL1 FIL2
R0E5 09/30/14 0013251 3 1 25 5 0 1 d d 56 0 1023 0001 510 511 7919 7919
R1E5 09/26/14 0013214 3 1 25 5 0 1 d d 45 0 1023 0001 514 509 7918 7918
R2E5 09/26/14 0013207 3 1 25 5 0 1 d d 30 0 1023 0001 511 506 7918 7918
R3E5 09/26/14 0013206 3 1 25 5 0 1 d d 32 0 1023 0001 510 507 7918 7918
R4E2 07/03/14 0012397 1 1 20 4 0 1 d d 42 1 10 6452 329 364 7658 7701
to rate and post a comment
Ygor Pereira
re: selenia aec error
October 03, 2014 10:13
Hello Juan, send me na e-mail with details, maybe I can help you
Regards
to rate and post a comment
Naelv
re: selenia aec error
October 07, 2014 05:59
the error 5 means "calculated exposure time less than a minimum" the exposure is too short. under what conditions this error occur? the AEC sensor is on automatic or it is on one of the positions? bye
to rate and post a comment
juan pablo amado
re: re: selenia aec error
October 08, 2014 05:34
I see in error log that AEC is satured (1023) then exp time is too small. AEC possition do not affect error frecuence. Fails when is in full auto mode