Over 150 Missouri Auctions End Today - Bid Now

seno 700 T problem

DJIELA Flavien

seno 700 T problem

July 20, 2009 09:00

Dear all,

I got a mammography equipment (seno 700T) which displays the following error code (E01, generator failure) during x ray exposure.
I have done the following tests:
- checking of the HV board ( every things is ok)
- replaced the generator ( the same problem occur)
When the error displays, by pressing on reset button on the CPU board it allows about five exams after what the same message is displayed again.

please, what else can i do

Best regards

Log inor Register

to rate and post a comment

(259)
(59)
(3)

Jose Vallejos

Re: seno 700 T problem

July 22, 2009 02:40

Dear DJiela,

You must change the x-ray tube, which is Gasified or Metalliced.

Anyway, try making Shots with low techniques . If no problems, increases the maximum techniques and try again some shots. If the problem occurs again, you know it.

Best Regards
Jose.

Log inor Register

to rate and post a comment

Fernando Canga

Re: seno 700 T problem

July 28, 2009 09:38

Dear,

You could try to perform a filament current calibration prior to change the Tube.

Regards

Fernando

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)