Over 500 Total Lots Up For Auction at One Location - NJ 06/15

senograph 600T

(13)
(1)

Emanuel Carmona

senograph 600T

December 01, 2005 09:29

Ihad problems with this system, when I prepare a shoot, the display shows E50 (error 50), but this error doesn't appear in the service manual. I hope that someone can help me with this problem.

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: senograph 600T

December 02, 2005 07:08

Emanuel:
Check the bucky, the error is for time out the bucky, disasembly the bucky module and check the motor and connection cable.

good luck

Log inor Register

to rate and post a comment

(18)

carlos andrioli

Re: senograph 600T

December 03, 2005 10:03

Check the fuse F7 in fuse painel. May be that it is broked.