Once the mamo is connected to the main supply the stop button iluminates. Since here everithing is ok. From time to time when press the ON button the light ON is switched on but nothing else happens.
I´ve checked all the voltages supplied from the power board, low voltage distribution board....
and it seems that all the voltages are present. I´ve checked a small DC converter 5V to +/- 15V
also.
Could somebody give me any clue?
That´s to those who participate in this fantastic forum
Re: senograph 800T problem in the start up sequence
October 06, 2007 10:14
Hi,
Did you check all emergency stitches. There are emergency switches in the tube stand also, other than in console. Check the main input power supply voltage settings also. need more details to support you further.
regards,
Shiju
Cura Medical Equipment Pvt Ltd
India.
shiju@cura.in
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
Daniel Fernandez
senograph 800T problem in the start up sequence
October 05, 2007 07:06
Once the mamo is connected to the main supply the stop button iluminates. Since here everithing is ok. From time to time when press the ON button the light ON is switched on but nothing else happens.
I´ve checked all the voltages supplied from the power board, low voltage distribution board....
and it seems that all the voltages are present. I´ve checked a small DC converter 5V to +/- 15V
also.
Could somebody give me any clue?
That´s to those who participate in this fantastic forum
to rate and post a comment
B Sahasrabuddhe
Re: senograph 800T problem in the start up sequence
October 06, 2007 04:24
Do u hear sounds of Bucky and tube itialisation?
Can you move compression paddle from keyboard?
Is the buzzer ON?
Wait for some time may be 3 min.
After this try to bring in Rotor. If rotor comes up try exposure. This will be blind shot.
Possible that your keyboard has gone. What type of keyborad do you have?
to rate and post a comment
SHIJU KARAKKOOTTIL
Re: senograph 800T problem in the start up sequence
October 06, 2007 10:14
Hi,
Did you check all emergency stitches. There are emergency switches in the tube stand also, other than in console. Check the main input power supply voltage settings also. need more details to support you further.
regards,
Shiju
Cura Medical Equipment Pvt Ltd
India.
shiju@cura.in
to rate and post a comment
Manuel Casanova
Re: senograph 800T problem in the start up sequence
October 10, 2007 07:40
Daniel :
Contact me for send you the service manual.
manuel