Senographe 600T hf error E 11

(158)
(30)
(1)

tahar nouri

Senographe 600T hf error E 11

June 27, 2008 04:08

hi everybody

I have a probleme with senographe 600T HF;displays E11 error , upon a power up sequence
relays K3; K2;K1 are enrgized only for a few seconds, after a while the relays are disengaged
and the error E11 is diplayed .

thank you in advance

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: Senographe 600T hf error E 11

June 28, 2008 06:02

Dear:

Yuo have a problem with STARTER.

Check the fuses.

Contact me at : semmedical@gmail.com for more explains.

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: Senographe 600T hf error E 11

June 30, 2008 12:00

Check:

1.- Fuse 18 in 201pl1
2.- Optocoupler B4 in 400PL7 GENERATOR INTERFACE BOARD.

Run the ANODE STARTER TEST JOb card LG003

I have parts.

semmedical@gmail.com

Log inor Register

to rate and post a comment

(158)
(30)
(1)

tahar nouri

Re: Senographe 600T hf error E 11

July 01, 2008 05:21

thank you very much
i have tested all boords on another good operating unit , all the boards (PL1 To Pl12)
are ok.
in addition to that , when i enrgize the K1 relay manually the KV window on control pannel is lit with no error, traying to make an exposure
E90 error message is displyed, knowing that relays on Starter board are working.

Log inor Register

to rate and post a comment

(1)

Marietta Salcedo

Re: Senographe 600T hf error E 11

July 23, 2009 01:14

I have a problem with the equipment, when I make an exposition the revelate is fine, but if I make a second exposition the revelate of the film is almost black. I assume that the KV is varying, but really I dont what's the cause.

Log inor Register

to rate and post a comment

(19)

(2)

VERNE CHAPPELL

Re: Senographe 600T hf error E 11

November 26, 2009 10:11

What was the fix for the E11 error.I have the same problem.Please let me know