March 02, 2013 03:01

Hi to all
i have a trouble with mammo 600T in compression ,
i think the problem is hydraulic cylinder parts
there is no oil . how add more oil to the this part.

i appreciate any help

xray.technician@yahoo.com

1 person liked this
Log inor Register

to rate and post a comment

(87)
(1)

roger dussart

re: Senographe 600T HF problem

March 02, 2013 09:19

there are no oil in system of compression.the system is pneumatic. Check tube plastics or sompressor membrane

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

re: Senographe 600T HF problem

March 05, 2013 11:40

I have the service manual email me crisb1955@aol.com

Log inor Register

to rate and post a comment

(22)
(1)

HUKAM CHAND

re: Senographe 600T HF problem

March 06, 2013 08:48

dear tahar

even if the oil leaks the compression should work. The oil in the 600t is just for auto compression release after pressing the paddle.

Is there any error coming on the machine display panel.
If no error is coming please check the mechanical board inside the machine. The board is located in the rack of the machine.

For anythinf please feel free to contact hukamsirhind@yahoo.com or aaahealthcare@gmail.com

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: Senographe 600T HF problem

March 06, 2013 09:59

check your email inbox