Hello, I have the following error in AOP/AEC mode(CNT/AUTO DOSE/AUTO STD/AUTO), E09 AEC failure 037/003, inhibiting the shot in any case, but i can shoot in manual mode with the following information: I20 037/003.
that can cause this failure?
you can check photocell cable because photocel'S cable can broke(ı meet), and refresh software so bıos battary is lowly may be empty on cpu board delete samething data .
error mean :
Go back to application mode
and note (FIX) the
erro code.This message is
often linked with error
037/003; invalid PM offset.
Perform AEC Photocell
checks for HVPM and PM
offsets (see Job Card )
E09 AEC Failure HTPM/IPM: offset value outside tolerance
Thanks for the answers.
what happens is that a colleague did the test to calibrate the photodetector, and the test obseleta stay the same, for this reason consulted. in the week I take the test and then I tell them.
Thanks.
hi, if someone has it you'll be grateful, this was replaced the photo-cell and HTPM board, what allowed me to do all calibrations, thank you all for your cooperation.
Regards.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Cristian Ganon
Senographe 700T E09 AEC failure
November 28, 2011 10:46
Hello, I have the following error in AOP/AEC mode(CNT/AUTO DOSE/AUTO STD/AUTO), E09 AEC failure 037/003, inhibiting the shot in any case, but i can shoot in manual mode with the following information: I20 037/003.
that can cause this failure?
to rate and post a comment
serkan yavaşça
Re: Senographe 700T E09 AEC failure
November 28, 2011 05:39
hi Cristian
ı dont meet thıs error but look it.
you can check photocell cable because photocel'S cable can broke(ı meet), and refresh software so bıos battary is lowly may be empty on cpu board delete samething data .
error mean :
Go back to application mode
and note (FIX) the
erro code.This message is
often linked with error
037/003; invalid PM offset.
Perform AEC Photocell
checks for HVPM and PM
offsets (see Job Card )
E09 AEC Failure HTPM/IPM: offset value outside tolerance
my email: serkanyvsc@hotmail.com
to rate and post a comment
Cristian Ganon
Re: Senographe 700T E09 AEC failure
November 29, 2011 08:05
Thanks for the answers.
what happens is that a colleague did the test to calibrate the photodetector, and the test obseleta stay the same, for this reason consulted. in the week I take the test and then I tell them.
Thanks.
to rate and post a comment
Cristian Ganon
Re: Senographe 700T E09 AEC failure
November 29, 2011 10:20
Hi, these are the tests I did:
CAL 004
SETUP/SETUP/AOP/CELL/ HV/DAC / FREQ/HV /CALIB/CALIB/START
Figure no "working" but I get "out of order"
it can be?
to try to detect the thickness of a plexi-glass, in thickness I figure "0" in the console.
wire HTPM is ok.
any recommendation is welcome.
regards.
to rate and post a comment
Jaime Diaz
Re: Senographe 700T E09 AEC failure
November 28, 2011 05:39
Probably you have the photocell damaged.
Regards
to rate and post a comment
Cristian Ganon
Re: Senographe 700T E09 AEC failure
March 01, 2012 08:38
hi, if someone has it you'll be grateful, this was replaced the photo-cell and HTPM board, what allowed me to do all calibrations, thank you all for your cooperation.
Regards.
to rate and post a comment
James McGinty
Re: Senographe 700T E09 AEC failure
March 02, 2012 10:15
if you have any more questions, we have some information that might help...
debi.breland@digitecmedical.com
to rate and post a comment