I have the following error in manual mode: E10 Heating Failure (022/006) and E5 Failure during exposure (079/002). In Automatic mode it fail when i use a dense material, but, for example it works with a pillow for phantom.
Thanks
Ralph, i check the senographe, and it seems to work fine in manual mode, but in automatic (CNT / Auto) the equipment dont shoot (it shoots (short shoot), then stop, again shoot, stop and so on until i depress the button).
I have to do the CAL 05?
Thanks
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Mario Silva
Senographe 700T Problem
April 08, 2009 09:21
I have the following error in manual mode: E10 Heating Failure (022/006) and E5 Failure during exposure (079/002). In Automatic mode it fail when i use a dense material, but, for example it works with a pillow for phantom.
Thanks
to rate and post a comment
Ralph Childs
Re: Senographe 700T Problem
April 08, 2009 02:55
Mario,
There is a very slight chance that it is the calibration. However, it is far more likely that the tube filament is beginning to fail.
Scooter
scanner1@eim.ae
to rate and post a comment
Mario Silva
Re: Senographe 700T Problem
April 13, 2009 01:30
Thanks Ralph. I will check the tube and get back to you.
to rate and post a comment
Mario Silva
Re: Senographe 700T Problem
April 18, 2009 09:29
Ralph, i check the senographe, and it seems to work fine in manual mode, but in automatic (CNT / Auto) the equipment dont shoot (it shoots (short shoot), then stop, again shoot, stop and so on until i depress the button).
I have to do the CAL 05?
Thanks
to rate and post a comment
catur suharinto
Re: Senographe 700T Problem
April 23, 2009 09:23
Unplug the HT cable, Fill the HT cable hole (in HT tank) with HT oil,replug the ht cable, then run mammocom.
try it