Hi.
I have a problem in a senographe, it is that in the films appears a dark strip in the chest side. I checked the filters, the grid, the films, cassettes and were ok.
It doesn't happen anytime.
Has anybody a clue?
Thanks.
to rate and post a comment
J.C. Navarro
Re: senographe 800T
September 17, 2007 03:13
Try to relocate the mouth marginator, (a metalic square window located abobe xray tube exit window)
If bad allocated it projects a shadow on the chest side of image.
to rate and post a comment
Daniel Fernandez
Re: senographe 800T
September 19, 2007 02:30
Thanks for your quick answer.
I will look over it.