Hi dotmed colleagues, I´m having a problem with a DMR, it wont power up, the stop light on the console won´t light, and the equipment can´t be turned on. already check for voltage on the 200 pl1 and it has -+8v and 2 green leds lighted. no other led is on in all the generator boards. can someone help me please.
Hello, Mario. Have you checked the Fuses? Since no other LED's are Lit, I would check Fuses F1 and F2. F1 is for the +12v and +12 v keep alive voltages. F2 is for the -8v.
Check the console connection in 400 PL1, check the F1 F2 Fuses, Check the physical condition of the 200 PL 1 Board.Also check the fused in 200 PL2.
If everything is OK, the problem with 200PL1.
Hi the problem was F1 on 200pl1. during boot there is another error 007/164 it says generator error. have you had this or know how to fix it? thanks regards
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
MARIO RAUL GONZALEZ
Senographe DMR
May 10, 2017 09:55
Hi dotmed colleagues, I´m having a problem with a DMR, it wont power up, the stop light on the console won´t light, and the equipment can´t be turned on. already check for voltage on the 200 pl1 and it has -+8v and 2 green leds lighted. no other led is on in all the generator boards. can someone help me please.
to rate and post a comment
Howard Karp
re: Senographe DMR
May 16, 2017 01:07
Hello, Mario. Have you checked the Fuses? Since no other LED's are Lit, I would check Fuses F1 and F2. F1 is for the +12v and +12 v keep alive voltages. F2 is for the -8v.
to rate and post a comment
MARIO RAUL GONZALEZ
re: re: Senographe DMR
May 17, 2017 05:23
thank you so much F1 on 200Pl1 was the problem
to rate and post a comment
Howard Karp
re: Senographe DMR
May 16, 2017 01:08
hello Mario. Have you checked the Fuses? Like F1 and F2? Since there are no other Leds Lit, I would start there.
to rate and post a comment
MARIO RAUL GONZALEZ
re: re: Senographe DMR
May 17, 2017 05:23
Hi Howard F1 was the problem
to rate and post a comment
Tecsila Service
re: Senographe DMR
May 17, 2017 10:28
Check the console connection in 400 PL1, check the F1 F2 Fuses, Check the physical condition of the 200 PL 1 Board.Also check the fused in 200 PL2.
If everything is OK, the problem with 200PL1.
to rate and post a comment
MARIO RAUL GONZALEZ
re: re: Senographe DMR
May 17, 2017 05:23
Hi the problem was F1 on 200pl1. during boot there is another error 007/164 it says generator error. have you had this or know how to fix it? thanks regards
to rate and post a comment
Tecsila Service
re: re: re: Senographe DMR
June 05, 2017 09:37
did you fix the problem?
to rate and post a comment