i have a problem with the system
it have a 6c2 transducer after a while when selecting the 6c2 probe system display error massage
"A hardware fault has occurred please call for service assistance"
after 2 days
when i power up the system with 6c2 connected it display the error massage any time i connect the probe the error massage displayed
i should remove it open system
i connect the probe to another seqouia machine it works well
any one can help????
do the other sequoia has the same software and pcb configuration? is it also a 512 or C512 share service? you may want to check the CN pcb and BF pcb rev in the system. Run diagnostics.
Hope it helps,
Rey
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
AHMEDELBAGA
seqouia 512
October 27, 2014 10:42
i have a problem with the system
it have a 6c2 transducer after a while when selecting the 6c2 probe system display error massage
"A hardware fault has occurred please call for service assistance"
after 2 days
when i power up the system with 6c2 connected it display the error massage any time i connect the probe the error massage displayed
i should remove it open system
i connect the probe to another seqouia machine it works well
any one can help????
to rate and post a comment
Rey Gonzalez
re: seqouia 512
October 27, 2014 06:30
do the other sequoia has the same software and pcb configuration? is it also a 512 or C512 share service? you may want to check the CN pcb and BF pcb rev in the system. Run diagnostics.
Hope it helps,
Rey
to rate and post a comment
AHMEDELBAGA
re: re: seqouia 512
October 28, 2014 09:47
the main machine C512 share service and probes was working well till this error ocured
the other one 512 GI where the probe works well