Sequoia Failed test :c004

Tags: Ultrasound, ACUSON

(30)
(6)
(1)

Charles Wollard

Sequoia Failed test :c004

July 16, 2015 08:15

I have a Sequoia 12.23 CRT system here, when running test suits 1 & 2 I am getting a failed test c004: pic adc vcc margin test error. I have replaced the PIC2 board with a know working one and still get the same error. Any ideas where to go from here?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Sequoia Failed test :c004

July 16, 2015 05:15

have you tried the power supply?

Log inor Register

to rate and post a comment

(30)
(6)
(1)

Charles Wollard

re: re: Sequoia Failed test :c004

July 16, 2015 06:19

I have replaced the PPS but not the main PS

Log inor Register

to rate and post a comment

(22)
(4)

Billy-Mac

re: Sequoia Failed test :c004

July 16, 2015 06:19

Charles. Put your original PIC2 back in and your Sequoia is good to go. The c004 error is caused by a backplane mod installed by Siemens. Apparently they cut a wire that supplies the pic adc vcc to cure some problem, but didn't update the diagnostics to ignore that the voltage wasn't there where the diagnostics is looking for it.

Log inor Register

to rate and post a comment

(30)
(6)
(1)

Charles Wollard

re: re: Sequoia Failed test :c004

July 17, 2015 09:52

Thank you, I wish I would have know this earlier it would have saved me a ton of aggravation. Thanks again!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)