Charles Wollard
Sequoia Failed test :c004
July 16, 2015 08:15
I have a Sequoia 12.23 CRT system here, when running test suits 1 & 2 I am getting a failed test c004: pic adc vcc margin test error. I have replaced the PIC2 board with a know working one and still get the same error. Any ideas where to go from here?
to rate and post a comment
Lenny Place
re: Sequoia Failed test :c004
July 16, 2015 05:15
have you tried the power supply?
to rate and post a comment
Charles Wollard
re: re: Sequoia Failed test :c004
July 16, 2015 06:19
I have replaced the PPS but not the main PS
to rate and post a comment
Billy-Mac
re: Sequoia Failed test :c004
July 16, 2015 06:19
Charles. Put your original PIC2 back in and your Sequoia is good to go. The c004 error is caused by a backplane mod installed by Siemens. Apparently they cut a wire that supplies the pic adc vcc to cure some problem, but didn't update the diagnostics to ignore that the voltage wasn't there where the diagnostics is looking for it.
to rate and post a comment
Charles Wollard
re: re: Sequoia Failed test :c004
July 17, 2015 09:52
Thank you, I wish I would have know this earlier it would have saved me a ton of aggravation. Thanks again!