Hello everyone.
I have a problem with neck coil neurovascular array (GE Signa Profile 0.2T). When scanning neck it's give artifact shading lower part. Please advise what to do. If will be need image I can send.
there is no substitute for seeing an image. Is it present in phantoms as well as humans? Is it present in a different coil using the same sequence? Is it repeatable? Is it in the same place always? Is it present with all sequences or just with some?
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
Doni Khasanov
Shading on neurovascular array coil
October 10, 2014 10:33
Hello everyone.
I have a problem with neck coil neurovascular array (GE Signa Profile 0.2T). When scanning neck it's give artifact shading lower part. Please advise what to do. If will be need image I can send.
to rate and post a comment
sford
re: Shading on neurovascular array coil
October 10, 2014 06:04
there is no substitute for seeing an image. Is it present in phantoms as well as humans? Is it present in a different coil using the same sequence? Is it repeatable? Is it in the same place always? Is it present with all sequences or just with some?
send the image to sford@proimagingservices.com
to rate and post a comment
Doni Khasanov
re: re: Shading on neurovascular array coil
October 11, 2014 02:06
on CTL coil images well. on phantom and humans present shading. and present in all sequence.
to rate and post a comment
Doni Khasanov
re: re: Shading on neurovascular array coil
October 11, 2014 02:06
on CTL coil images well. on phantom and humans present shading. and present in all sequence.
to rate and post a comment
Doni Khasanov
re: Shading on neurovascular array coil
October 11, 2014 02:06
on CTL coil images well. on phantom and humans present shading. and present in all sequence.
to rate and post a comment