(1) power down the systems, pull out the table manulay to full out to activate the out limit switch, then check the status.
(2) check the panel scanner i/f board in console and its connectors on its backside on backplance.
(3)check for any loose connector on scan cont. board and motor driver board in gantry
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
Charles Psarreas
shimadzu 5000
May 27, 2004 05:00
scanner came up in the morning, blicking 505 and the bed would not go in or out.
check all power supplies ok.
gantry tilts ok - but will not tilt +
can not find 505 in manual , could this be a sos for a error?
anyone have a idea?
charlie Psarreas
516 250 6292
to rate and post a comment
Iftikhar Khan
Re: shimadzu 5000
June 03, 2004 01:13
(1) power down the systems, pull out the table manulay to full out to activate the out limit switch, then check the status.
(2) check the panel scanner i/f board in console and its connectors on its backside on backplance.
(3)check for any loose connector on scan cont. board and motor driver board in gantry
wish u all the best
iftikhar
+923037869500
to rate and post a comment
Carlos Duran
Re: shimadzu 5000
June 18, 2004 05:45
hello:
you can see the trouble via software, with plnchk in CLI mode, to view if the data goes well between console and gantry.
good luck
to rate and post a comment