Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

siemans somatom 16 slice ct table problem

Tags: CT, SIEMENS

(5)
(1)

kenneth shives

siemans somatom 16 slice ct table problem

March 20, 2013 03:20

iam working on a siemans somatom 16 slice ct and the table is all the way up and the problem is when you press the down botton you can hear a relay come in then nothing it like shuts down and then comes back on table stuck can someone plesae help thankyou

1 person liked this
Log inor Register

to rate and post a comment

(8)
(1)

Bill Schimmer

re: siemans somatom 16 slice ct table problem

March 20, 2013 09:04

There should be an access panel at the bottom of the table near the gantry on the pt right side. Inside is an override so you can lower the table to middle travel. Check your foot switches for proper operation and end limit switches.

1 person liked this
Log inor Register

to rate and post a comment

(5)
(1)

kenneth shives

re: re: siemans somatom 16 slice ct table problem

March 22, 2013 10:08

i really do thank you for the input.i checked the limit switch its good,you know that long worm gear or screw shaft the motor turns to drive table ud or down.well the the non motor side has that maglock,the 4 ajusting bolts backed them off 1 half turn .its started working but it was still having a hard time so i backed them a quarter turn more its working fine now do you happen to know if there supose to be pouer to that at all times there were no drawing for me to check cause there is voltage that never drops out.and thanks again getting back.thankyou ken

Log inor Register

to rate and post a comment

saju

re: re: re: siemans somatom 16 slice ct table problem

March 22, 2013 05:04

put grease on the shaft and do two oe three times fully up and down from control panel.may be the grease has gone dry

Log inor Register

to rate and post a comment

(8)
(1)

Bill Schimmer

re: siemans somatom 16 slice ct table problem

March 20, 2013 09:05

There is an access panel on the table at the patients right side near the gantry. Put the table into override and check the operation of the footswitch and end limit switches.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: siemans somatom 16 slice ct table problem

March 21, 2013 10:14

what is error code?
if you don't have service password, just send me the Dougle #, I will send you the password, then you can look at the log file

Log inor Register

to rate and post a comment

saju

re: siemans somatom 16 slice ct table problem

March 21, 2013 03:58

if its a watercooled system . below table there is shaft from motor you can put a spanner and make it mechanically down before that see whether you have transport block fixed .if not you can try this,

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: siemans somatom 16 slice ct table problem

March 21, 2013 09:38

you may need to change D702 board(table moving controler)