Siemens 300 ventilator problem

(11)

Kaushal Patil

Siemens 300 ventilator problem

March 24, 2008 09:49

Hi,

I am facing problem with Siemens 300 ventilator your views,comment are welcomed.

The siemens 300 ventilator shows much lesser value in display panel of INSP tidal volume (ONLY in NEONATAL mode) but EXP tidal volume matches with set INSP tidal volume.

I have changed Air module but was not of any help.

Can you help in diagnosing the problem?

Thanks,

Kaushal Patil.

Log inor Register

to rate and post a comment

(76)
(4)
(34)

Lakshmi Narayana

Re: Siemens 300 ventilator problem

March 24, 2008 01:48

In which mode this is happening?

Log inor Register

to rate and post a comment

(11)

Kaushal Patil

Re: Siemens 300 ventilator problem

March 25, 2008 12:18

Hi,
Volume and Pressure, both modes. I had similar problem but on Exp Tidal volume reading which was solved by changing Air Module.
Now it is in Insp Tidal Volume in Neonatal range in Volume and Pressure mode.

Thanks for your reply.

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: Siemens 300 ventilator problem

March 25, 2008 02:16

Hi,
Please check if cascade humidifier / patient circuit has the appropiate size. Incorrect size may cause the death volume be to high for the patient size and produce this symptom just in neonatal mode. Best regards
José

Log inor Register

to rate and post a comment