I have a Siemens Antares Ultrasound
System run good without probes are connected
but, when probes are connected the system turn off after 15 min run.
any idea which will bee the problem?
Thanks for all you help

1 person liked this
Log inor Register

to rate and post a comment

(3)

Kemal Demir

re: Siemens Antares Problem

July 02, 2016 02:50

First you need to check Power supply maybe its have dust clean it but remove the system then open the power supply inside clean then reverse order the job is finish check the machine