Siemens Arcadis Orbic errors 5901 6573

Tags: C-Arm, SIEMENS

(67)
(3)
(1)

Lobsang Ortiz

Siemens Arcadis Orbic errors 5901 6573

May 30, 2023 01:53

Hello everybody,

I have trouble with a Siemens Arcadis Orbic 3D, error E5901, and parameter out of range ASP_MSY_E_6573, I replaced the x ray tube, but the error still present, the manual says that could be D1 or D2 board, I replaced D2 board but this is not the problem, so I try to replace the D1 board and now appears The main system is disconnected, so I have no comunication between trolley and C arm, I don´t know if the D1 board is plug and play or if I can do something else, I can´t load the software to de C arm because It appears disconnected and in the service it´s not appear the main system menu, What can I do?

Thanks in advance.

Lobsang Ortiz

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: Siemens Arcadis Orbic errors 5901 6573

June 01, 2023 10:41

From long time ago, you can't just plain exchange boards on C arms. (This is valid for most manufacturers).

There's a tasks you must do to put in working order the replacement.