Hi Miguel,
Error 2070
Explanation: a) During power up, ADC tests fails
Action: a1) Check the voltage supply M14 (+/-15V)
a2) If the fault doesn't come if the plug D1 X1 is disconected
please check the connection to the TV iris.
a3) Check cables from voltage supply M14
to board D1 Main System
a4) The Board M14 might be defect.
a5) The Main System D1 might be defect.
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Miguel Chavira
siemens arcadis varic error
April 07, 2021 10:09
hi, i have a troble with a c arm siemens arcadis varic
During power up ADC tests failed
if the failure occurs frequently , inform service
Error 2070
if somebody knows about this error i will aprecciate the support.
miguel chavira
to rate and post a comment
Marko Shevchuk
re: siemens arcadis varic error
April 08, 2021 10:42
Hi Miguel,
Error 2070
Explanation: a) During power up, ADC tests fails
Action: a1) Check the voltage supply M14 (+/-15V)
a2) If the fault doesn't come if the plug D1 X1 is disconected
please check the connection to the TV iris.
a3) Check cables from voltage supply M14
to board D1 Main System
a4) The Board M14 might be defect.
a5) The Main System D1 might be defect.
to rate and post a comment
Miguel Chavira
re: re: siemens arcadis varic error
April 09, 2021 01:38
i will do
Maybe will be the D1 board
Because i habe checked all situations
Thank’s so much
to rate and post a comment
Sergey Zheleznyak
re: siemens arcadis varic error
April 08, 2021 10:54
Hello, my Friend!
Did you enter service browser and read error description in error log?
to rate and post a comment
Aleksandr Chebot
re: siemens arcadis varic error
April 08, 2021 10:55
Hi! You must go to service and see errors. Write me on my mail if need help with service
to rate and post a comment