Error E8515 = Videomed DC -9V out of range
Possible trouble: Videomed DC -9V power supply for analog circuits
outside the valid tolerance range.
Possible help: Check power supply of the Videomed DC. Videomed DC defective.
Error E2005 = Videomed DC, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC. Possibly D4 or D1 defective.
Error 8515
Possible trouble: Videomed DC -9V power supply for analog circuits
outside the valid tolerance range.
Possible help: Check power supply of the Videomed DC.
Videomed DC defective.
Error 2005
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC.
Check service switches on D4.
In normal use, all switches s1 to S8 should be set to OFF.
Possibly D4 or D1 defective.
Error E8515 Videomed DC -9V out of range
Possible trouble: Videomed DC -9V power supply for analog circuits
outside the valid tolerance range.
Possible help: Check power supply of the Videomed DC.
Videomed DC defective.
Error E2005 Videomed DC, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC.
Possibly D4 or D1 defective.
Error E2005 Videomed DC, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC.
Possibly D4 or D1 defective.
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
ZAFAR ULLAH
Siemens Compact C arm
April 18, 2013 11:55
Hi there,
We are having Error 8515 and error 2005 on Siemens Compact C Arm.
Can anybody kindly help us to find out the possible cause
thanks
Zafar
to rate and post a comment
Wayne Horsman
re: Siemens Compact C arm
April 19, 2013 09:13
Error E8515 = Videomed DC -9V out of range
Possible trouble: Videomed DC -9V power supply for analog circuits
outside the valid tolerance range.
Possible help: Check power supply of the Videomed DC. Videomed DC defective.
to rate and post a comment
Wayne Horsman
re: Siemens Compact C arm
April 19, 2013 09:14
Error E2005 = Videomed DC, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC. Possibly D4 or D1 defective.
to rate and post a comment
Ralph Frizzle
re: Siemens Compact C arm
April 19, 2013 09:16
Error 8515
Possible trouble: Videomed DC -9V power supply for analog circuits
outside the valid tolerance range.
Possible help: Check power supply of the Videomed DC.
Videomed DC defective.
Error 2005
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC.
Check service switches on D4.
In normal use, all switches s1 to S8 should be set to OFF.
Possibly D4 or D1 defective.
to rate and post a comment
ARIAS RODRIGUEZ JOSE
re: Siemens Compact C arm
April 19, 2013 09:16
Error E8515 Videomed DC -9V out of range
Possible trouble: Videomed DC -9V power supply for analog circuits
outside the valid tolerance range.
Possible help: Check power supply of the Videomed DC.
Videomed DC defective.
Error E2005 Videomed DC, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC.
Possibly D4 or D1 defective.
to rate and post a comment
ARIAS RODRIGUEZ JOSE
re: Siemens Compact C arm
April 19, 2013 09:17
Error E2005 Videomed DC, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Videomed DC.
Possibly D4 or D1 defective.
to rate and post a comment