Siemens Concerto Error

(98)
(15)
(15)

Carlos Lopez

Siemens Concerto Error

January 21, 2011 10:36

The following message appears on the monitor of a Siemens Concerto Open MRI.

Frequency adjustment error. Received signal is too low. Check patient positioning/coil selection/adjustment volume.

Any ideas?

Log inor Register

to rate and post a comment

(34)
(6)

Dmitry Ogosyn

Re: Siemens Concerto Error

January 23, 2011 09:12

Frequency adjustment error. Received signal is too low. Check patient positioning/coil selection/adjustment volume.

Does It is on all coils? If not, connect the body large coil. In brass "system" disable one coil plag, X1 for example, system is run? If not, activate X1, disable X2. System run? If yes, defective coil element ore RF path (may be BTB). Check another coil. If system does not run, check the temperate of magnet (app. 31,5 degre). If Ok, use manual adjustment , for activate this menu, any protocol
Must be open. Check isocentr shim in seso platform

Log inor Register

to rate and post a comment

(1)
(1)

Ronnie Licariao Aguiar

re: Re: Siemens Concerto Error

February 24, 2014 02:40

please send me you mail,my is imageassistance@uol.com.br

1 person liked this
Log inor Register

to rate and post a comment

(34)
(6)

Dmitry Ogosyn

re: re: Re: Siemens Concerto Error

March 04, 2014 03:26

OK

Log inor Register

to rate and post a comment

(34)
(6)

Dmitry Ogosyn

re: re: Re: Siemens Concerto Error

March 04, 2014 03:26

I sent you my e mail/

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We repair MRI Coils, RF amplifiers, Gradient Amplifiers and Injectors.

MIT labs, experts in Multi-Vendor component level repair of: MRI Coils, RF amplifiers, Gradient Amplifiers Contrast Media Injectors. System repairs, sub-assembly repairs, component level repairs, refurbish/calibrate. info@mitlabsusa.com/+1 (305) 470-8013

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)