Carlos Lopez
Siemens Concerto Error
January 21, 2011 10:36
The following message appears on the monitor of a Siemens Concerto Open MRI.
Frequency adjustment error. Received signal is too low. Check patient positioning/coil selection/adjustment volume.
Any ideas?
to rate and post a comment
Dmitry Ogosyn
Re: Siemens Concerto Error
January 23, 2011 09:12
Frequency adjustment error. Received signal is too low. Check patient positioning/coil selection/adjustment volume.
Does It is on all coils? If not, connect the body large coil. In brass "system" disable one coil plag, X1 for example, system is run? If not, activate X1, disable X2. System run? If yes, defective coil element ore RF path (may be BTB). Check another coil. If system does not run, check the temperate of magnet (app. 31,5 degre). If Ok, use manual adjustment , for activate this menu, any protocol
Must be open. Check isocentr shim in seso platform
to rate and post a comment
Ronnie Licariao Aguiar
re: Re: Siemens Concerto Error
February 24, 2014 02:40
please send me you mail,my is imageassistance@uol.com.br
to rate and post a comment
Dmitry Ogosyn
re: re: Re: Siemens Concerto Error
March 04, 2014 03:26
OK
to rate and post a comment
Dmitry Ogosyn
re: re: Re: Siemens Concerto Error
March 04, 2014 03:26
I sent you my e mail/
to rate and post a comment