Siemens e.cam motion problem

(2)

Ilyan Dukov

Siemens e.cam motion problem

May 13, 2018 03:18

A motion problem occurred.
The error shown when trying to run an analysis is:
“The camera reported an error:A motion control server error has occurred (1023:100)” on the Syngo monitor.And the analysis is aborted.
The problem does not occur when moving the patient bed or Gantry manually.Nor when performing home position.
The problem occurs only when moving the patient bed horizontally, during long analysis, e.g. bones, iodine 131.
The problem disappears if the analysis is performed in the bottom end of the patient bed.
The Syngo software is MI_VA46C.SNAC for VA46C, version 8.7.10.6 QNX 6.3.
I’ve reinstalled SNAC on a new HDD in order to eliminate the possibility of a software problem or a problem reading the HDD.
But the problem persists.
Could you please share some advice or an insight as to what may be causing this problem and how I could solve it?
Siemens Bulgaria won’t cooperate.
Thank you in advance.

Log inor Register

to rate and post a comment

Eric Hauk

re: Siemens e.cam motion problem

May 14, 2018 09:56

How far are you from Banja Luka? I have a friend who could come over and repair it. It’s likely just a potentiometer or encoder for the motor.

Log inor Register

to rate and post a comment

Eric Hauk

re: Siemens e.cam motion problem

May 14, 2018 09:56

Oh forgot if you want me to put you in contact with the guy in Eastern Europe my email is ehaukimaging@gmail.com

Log inor Register

to rate and post a comment

(14)
(1)

Pablo Piccinini

re: Siemens e.cam motion problem

May 14, 2018 04:19

Dear Ilyan, could you send your email? I will keep in touch with you about the problem.
Regards

Pablo

Log inor Register

to rate and post a comment

(2)

Ilyan Dukov

re: re: Siemens e.cam motion problem

May 17, 2018 09:16

Hi Pablo my email is ack.ltd@gmail.com

Log inor Register

to rate and post a comment