Hello! I have a problem with siemens ecam. After stopping for 2 months, after switching on, the problem "both the detectors are not initialized correctly" appeared. The second detector does not turn on and there is no voltage on it. The fan inside the detector does not rotate. At the input of the detector power supply, the voltage is 105 volts. What is the reason?
Check the input voltage on the other detector if at 105 you may want to retap the cleanbox and or change it. Your OEM power supply is blown in head 2 or the fuse is blown. Check the 15amp tab fuse in the head of good replace the supply.
You have a problem in Detector 2, then you have to focus just in detector 2. The communication between Snac and detector is serie in this way: Snac - Detector 2 - Detector 1. Then If you had a problem in detector 2, you woudnt have communication with detector 1.
You have to check in detector 2, the power supply, might this power supply is not working. 105 Voltage is ok.
I am sure your problem is detector 2 power supply.
Best regards
Problem solved. Malfunction of the standby unit on the chip UC3844. One of the 10 Ohm resistors in the transistor source circuit is faulty. The standby unit could not start due to overcurrent. It works! Thanks everyone!
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Egor Shamov
Siemens ecam problem detectors
September 17, 2019 10:39
Hello! I have a problem with siemens ecam. After stopping for 2 months, after switching on, the problem "both the detectors are not initialized correctly" appeared. The second detector does not turn on and there is no voltage on it. The fan inside the detector does not rotate. At the input of the detector power supply, the voltage is 105 volts. What is the reason?
to rate and post a comment
Eric Hauk
re: Siemens ecam problem detectors
September 18, 2019 10:29
Check the input voltage on the other detector if at 105 you may want to retap the cleanbox and or change it. Your OEM power supply is blown in head 2 or the fuse is blown. Check the 15amp tab fuse in the head of good replace the supply.
to rate and post a comment
Jorge Barrios
re: Siemens ecam problem detectors
September 18, 2019 10:32
Hello Egor,
You have a problem in Detector 2, then you have to focus just in detector 2. The communication between Snac and detector is serie in this way: Snac - Detector 2 - Detector 1. Then If you had a problem in detector 2, you woudnt have communication with detector 1.
You have to check in detector 2, the power supply, might this power supply is not working. 105 Voltage is ok.
I am sure your problem is detector 2 power supply.
Best regards
to rate and post a comment
Egor Shamov
re: re: Siemens ecam problem detectors
September 19, 2019 09:59
Problem solved. Malfunction of the standby unit on the chip UC3844. One of the 10 Ohm resistors in the transistor source circuit is faulty. The standby unit could not start due to overcurrent. It works! Thanks everyone!