I am getting an Error on a Siemens SX 65/80. It is 715/080. Does anyone have experience with this error? Furthermore, I am looking for the software and a service manual. I heard the software will only on older Microsoft OS.
Error 715 0
Meaning:
Minus-kVmin error
Description:
The minus kVmin monitoring becomes active if the -kV actual value is more than 10 kV
less than the -kV nominal value.
Display LED D100.V165 "kV MIN-"
Check both ends of the negative HT cable for signs of arcing, I have also had this problem in the past and the HT transformer itself was faulty.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
MedicalTao
Siemens Error 715/080
January 03, 2019 12:04
I am getting an Error on a Siemens SX 65/80. It is 715/080. Does anyone have experience with this error? Furthermore, I am looking for the software and a service manual. I heard the software will only on older Microsoft OS.
to rate and post a comment
mahdi Bennai
re: Siemens Error 715/080
January 04, 2019 09:47
Hi. I can help
contact.medicotech@gmail.com
to rate and post a comment
Lister
re: Siemens Error 715/080
January 04, 2019 09:47
Error 715 0
Meaning:
Minus-kVmin error
Description:
The minus kVmin monitoring becomes active if the -kV actual value is more than 10 kV
less than the -kV nominal value.
Display LED D100.V165 "kV MIN-"
Check both ends of the negative HT cable for signs of arcing, I have also had this problem in the past and the HT transformer itself was faulty.