Error E7151 Slot diaphragm rotation 0 degr. pos. not found
ossible trouble: The slot diaphragm rotation 0 degree position
was not found during initialization.
Possible help: Check the mechanical movement.
Check the motor.
Check the motor wires.
Check the actual value potentiometer wires.
Check the actual value potentiometer.
Board D1 defective.
In my experience, it is usually a stripped lower ring gear in the slot diaphragm portion of the collimator assembly. Siemens does not sell this gear, just the whole collimator assembly. Call me, or email me, with the part number of your collimator (might be 2829880). We can probably do a "repair & return" for you.
Possible trouble:
The slot diaphragm rotation 0 degree position was not found during initialization.
Possible help:
Check the mechanical movement.
Check the motor.
Check the motor wires.
Check the actual value potentiometer wires.
Check the actual value potentiometer.
Board D1 defective.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Gary Benitez
Siemens Iso-C error code 7151
January 19, 2015 05:08
Does anyone know what error code 7151 is on a Siemens Iso-c?
to rate and post a comment
Lenny Place
re: Siemens Iso-C error code 7151
January 21, 2015 09:47
Error E7151 Slot diaphragm rotation 0 degr. pos. not found
ossible trouble: The slot diaphragm rotation 0 degree position
was not found during initialization.
Possible help: Check the mechanical movement.
Check the motor.
Check the motor wires.
Check the actual value potentiometer wires.
Check the actual value potentiometer.
Board D1 defective.
to rate and post a comment
Wayne Horsman
re: Siemens Iso-C error code 7151
January 21, 2015 09:47
Gary -
In my experience, it is usually a stripped lower ring gear in the slot diaphragm portion of the collimator assembly. Siemens does not sell this gear, just the whole collimator assembly. Call me, or email me, with the part number of your collimator (might be 2829880). We can probably do a "repair & return" for you.
- Wayne H.
======= Error E7151 =====
E7151 => Slot diaphragm rotation 0 degr. pos. not found.
Possible trouble:
The slot diaphragm rotation 0 degree position was not found during initialization.
Possible help:
Check the mechanical movement.
Check the motor.
Check the motor wires.
Check the actual value potentiometer wires.
Check the actual value potentiometer.
Board D1 defective.