I have mobilett plus e system which has error message 6 when I make preparation
Can you inform me what is this message means and how can I solve this problem
Thank you

Log inor Register

to rate and post a comment

DMEDS

re: siemens mobilett plus e problem

October 04, 2017 09:24

Dear Safwan,
I have all manuals for this system ,
if you need it contact me,
magmesme@gmail.com
Error 6 meaning:-
During preparation or exposure
The rotating anode
frequency cannot be maintained above 142 Hz.
*At preparation: The anode is accelerated
for 2 s, the speed is measured.
If the rotation is too slow, the acceleration
phase is repeated and retested.
If the rotation is still too slow or the
anode looses speed too fast, a message
will be displayed.
Remove K64 page 140** during the
next test.

1 person liked this
Log inor Register

to rate and post a comment

DMEDS

re: siemens mobilett plus e problem

October 04, 2017 09:24

• Activate the radiation release
switch to the first level
(preparation) and listen for the
anode to start up; refer to page
150**.
• Check that the following supply
voltages are present on D4,
+100 V, +15 V, +5 V DC, refer to
page 150**.

1 person liked this
Log inor Register

to rate and post a comment

DMEDS

re: siemens mobilett plus e problem

October 04, 2017 09:24

• Check that the trigger pulses are
present at "ROT"; when the anode
is accelerated, refer to page 150**.
• Check that the measuring signals
"ROT_VAL" on D4 and "ROT" on
D1 are OK, refer to page 150**.

1 person liked this
Log inor Register

to rate and post a comment

DMEDS

re: siemens mobilett plus e problem

October 04, 2017 09:24

• Turn off the unit and remove K61.
Measure the stator winding in the
power plug.
➪ K61: 3-12 = 8
➪ K61: 1-2 = 5

1 person liked this
Log inor Register

to rate and post a comment

(3)

Safwan Monajed

re: re: siemens mobilett plus e problem

October 05, 2017 09:05

Thanks for all your advices