siemens mobilett plus err 21

February 12, 2025 03:39

Dear Friends,
I have been facing err 21 on siemens mobilett plus. (The measured value of the standby filament current has
exceeded the limit.)
D4, +60 V, +15 V and +5 V are normal.
"FIL_VAL" on D1 must be between
0.16 and 2.5 V when the filament is
in standby mode. but mine is 12,5 volt.
Could you please assist me to solve the matter? Thank you

2 people liked this
Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: siemens mobilett plus err 21

February 18, 2025 10:28

Before suspecting failure in H1, check cable integrity (short circuit) from D4 ( filament) to H1 (HV tank) ; open the filament circuit at Fil the connector on H1 , I sense must go to cero. Refer to wiring diagram document.
If not check D4 or cable broken

Log inor Register

to rate and post a comment

(26)
(2)

lei wang

re: siemens mobilett plus err 21

March 05, 2025 01:05

Remove the connection line between the filament plate and the combined ball tube, and turn on the computer to see if there is any change in the error.

Log inor Register

to rate and post a comment

(26)
(2)

lei wang

re: siemens mobilett plus err 21

March 05, 2025 01:05

Remove the connection line between the filament plate and the combined ball tube, and turn on the computer to see if there is any change in the error.

Log inor Register

to rate and post a comment