Hello The team works by charging capacitors, as have the 110 Volts AC supply there should be no problem.
Information herror command you, you need to diagrams to work, I have a part but is incomplete.
I hope you serve and you solve your problem, my email to keep track of your progress and can be contacted as follows floresr_kco@hotmail.com
Err 9 During exposure When measuring the
kV value, a large high tension asymmetry
is detected.
Normally the kV value is symmetric around ground. The acceptable deviation is ± 19.5 kV. This corresponds
to HTD 2.5 V± 0.5 V.
· Intermittent measurement cables or defective single tank.
· Check the the measurement cable K9 between D1 and the single tank.
· Check the current to the single tank
· Measure the HTD signal on D1 and compare it with notes and curves.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
mdastech
Siemens Mobilett Plus: Err 9
January 20, 2016 01:41
Hi All,
We just installed a xray in a remote hospital. Whenever the X-ray is fired, it gives an Err 9. Does anyone know what might be causing this?
More info: The Xray machine is connected to a 5000 watts transformer to convert from 220 to 110V...could this be the problem?
Any help will be appreciated...thanks.
to rate and post a comment
CARLOS FLORES
re: Siemens Mobilett Plus: Err 9
January 21, 2016 10:17
Hello The team works by charging capacitors, as have the 110 Volts AC supply there should be no problem.
Information herror command you, you need to diagrams to work, I have a part but is incomplete.
I hope you serve and you solve your problem, my email to keep track of your progress and can be contacted as follows floresr_kco@hotmail.com
Err 9 During exposure When measuring the
kV value, a large high tension asymmetry
is detected.
Normally the kV value is symmetric around ground. The acceptable deviation is ± 19.5 kV. This corresponds
to HTD 2.5 V± 0.5 V.
· Intermittent measurement cables or defective single tank.
· Check the the measurement cable K9 between D1 and the single tank.
· Check the current to the single tank
· Measure the HTD signal on D1 and compare it with notes and curves.
to rate and post a comment