Standby Declaration: D916 detects actual kV error.
Action:
ยท Check the cable connection of measurement data log D907, the single tank, and D916.X8 (CPU).
Possible causes: Defective single tank, defective cable, defective D916 (CPU).
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Ricardo Paulin
siemens mobilette error code 8
May 05, 2007 06:01
Does anybody knows what is the error code 8, in a Siemens Mobilette?
Thankyou
to rate and post a comment
Henrique Gomes
Re: siemens mobilette error code 8
May 07, 2007 09:21
If you have the Mobilett XP:
ERR 8 Static kV measured value 0
System mode:
Standby Declaration: D916 detects actual kV error.
Action:
ยท Check the cable connection of measurement data log D907, the single tank, and D916.X8 (CPU).
Possible causes: Defective single tank, defective cable, defective D916 (CPU).
Regards
Gomes.