Over 150 Missouri Auctions End Today - Bid Now

Siemens Multix Compact K

Tags: Radiology - R/F, SIEMENS

(19)

Hassan Elkhatib

Siemens Multix Compact K

January 19, 2018 09:09

The device lost calibration, the equipment was calibrated but not the AEC (maybe the boards D500 and D300 are damaged)?
Actaully the AEC mode doesn't work.
Now it is handled with the device manually without AEC.

Please let me know a solution.

Thank you!

Log inor Register

to rate and post a comment

DMEDS

re: Siemens Multix Compact K

January 19, 2018 03:28

Hello,
i have the manuals for system
Multix-CH
and
Multix-CPH
if you need any contact me,
magmesme@gmail.com
Regards
thanks