Siemens Multix IT table bucky problem

Ron Rowley

Siemens Multix IT table bucky problem

June 16, 2012 11:34

Siemens Multix IT. The collimator flashes from Manual to ACSS when trying to do table bucky exposures. Will not expose or collimate to table bucky. Works properly when the tube is rotated towards the wall bucky. This is an intermittant problem.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: Siemens Multix IT table bucky problem

June 17, 2012 05:51

Looks like a wire broken inside cable connector at collimator unit.
Move the cable yourself to try repeat the failure, check the pins at connector too.

Good luck!

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Siemens Multix IT table bucky problem

June 17, 2012 05:51

Check SID sensor or cassette size sensing circuit

Log inor Register

to rate and post a comment

Fernando Canga

re: Siemens Multix IT table bucky problem

June 18, 2012 09:20

Try to clean the detector photodiodes. You must remove the table, the ion chambers, the plate that cover the photodiodes and clean them. Normally it works.

Check also that the cassete is placed correctly over the tray.

Regards

Log inor Register

to rate and post a comment