- "file read error" file not found or version mismatch. Do calibration or set fix defect pixel file.
- "error of detector (-5101)"
- "fatal error" .. ccrstoreEnqueue
- "fatal error" .. ccrqEnqStudy-1
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
MIGUEL ANGEL TORRES MARTINEZ
Siemens multix MT
November 05, 2015 05:57
Someone, Could you help me, please?
Show error message:
- "file read error" file not found or version mismatch. Do calibration or set fix defect pixel file.
- "error of detector (-5101)"
- "fatal error" .. ccrstoreEnqueue
- "fatal error" .. ccrqEnqStudy-1
Thanks
(I speak spanish)
to rate and post a comment
Lenny Place
re: Siemens multix MT
November 11, 2015 10:24
probably a bad detector cable.
-5101
Detector error
Possible error source:
· The cable is not connected even though it is adjusted.
Corrective action:
· Check the detector cable connection. If you want to recover the system, shut it down.