Hi
This error indicates that there is something wrong with K2 charger contactor.
- Check whether the K2 breaker is active during switch-on; if needed, replace it.
- Check, and if needed, replace the cables between D501 X5 and the charger contactor
(W504).
Err023: K2 Charger Contactor
Description:
· An error with the K2 charger contactor was detected.
Possible Causes of Error:
· K2 charger contactor (charge breaker) defective
· Cabling between D501 and charger contactor defective.
Corrective Measures:
· Check whether the K2 breaker is active during switch-on; if needed, replace it.
· Check, and if needed, replace the cables between D501 X5 and the charger contactor (W504).
We also have a swing with error 001.
It seems like a problem with main power supply but it looks everything ok. Do you have any idea what might be the problem?
Thank you in advance.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
laric
siemens multix swing error 0023
April 08, 2013 03:45
i've error 0023 on my siemens multix swing.
MODEL 09815770
please can anybody tell me the meaning of that error and how to resulve it ?
yours sincerely
larictameye@yahoo.fr
to rate and post a comment
ghazy
re: siemens multix swing error 0023
April 09, 2013 09:18
Hi
This error indicates that there is something wrong with K2 charger contactor.
- Check whether the K2 breaker is active during switch-on; if needed, replace it.
- Check, and if needed, replace the cables between D501 X5 and the charger contactor
(W504).
to rate and post a comment
xiaohua xu
re: siemens multix swing error 0023
April 13, 2013 11:36
Err023: K2 Charger Contactor
Description:
· An error with the K2 charger contactor was detected.
Possible Causes of Error:
· K2 charger contactor (charge breaker) defective
· Cabling between D501 and charger contactor defective.
Corrective Measures:
· Check whether the K2 breaker is active during switch-on; if needed, replace it.
· Check, and if needed, replace the cables between D501 X5 and the charger contactor (W504).
to rate and post a comment
ALEXANDER PILAFTSIS
re: re: siemens multix swing error 0023
April 17, 2014 08:24
We also have a swing with error 001.
It seems like a problem with main power supply but it looks everything ok. Do you have any idea what might be the problem?
Thank you in advance.
to rate and post a comment
Carlos Jaime
re: siemens multix swing error 0023
June 17, 2014 03:53
Hello i looking for the manual for siemens multix swing cause I have 0038 y 0041 erros on the display,can you help me??
to rate and post a comment
Carlos Jaime
re: siemens multix swing error 0023
December 10, 2014 08:43
i've error 0030 y 0032 on my siemens multix swing, can you help me to resulve it?
to rate and post a comment