SIEMENS MULTIX SWING POLYDROS IT

December 12, 2016 04:40

I HAVE A SIEMENS MULTIX SWING POLYDROS IT X-RAY MACHINE WITH ERROR 631.

I HAVE CHECKED THE TUBE ROTOR, IT SEEMS TO BE RUNNING WELL. ANY ASSISTANCE?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: SIEMENS MULTIX SWING POLYDROS IT

December 13, 2016 10:10

This is a Tomo issue not a tube rotor issue check for a blown fuse on the D54 board. If you need any more info please send me an email address


Tomo drive. Control error ==> tube rotation motor movement to hard or defective or D54 defective

Log inor Register

to rate and post a comment

(121)
(33)
(9)

joe ssali

re: re: SIEMENS MULTIX SWING POLYDROS IT

December 14, 2016 10:31

Thank you Lenny. Very grateful. Do you have the service manual? And do you have access to parts for this machine.

Log inor Register

to rate and post a comment

(121)
(33)
(9)

joe ssali

re: SIEMENS MULTIX SWING POLYDROS IT

December 15, 2016 10:19

Thank you Lenny and Sanchez. The Machine is working now. All because of your support

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: SIEMENS MULTIX SWING POLYDROS IT

December 15, 2016 02:59

I am glad you were able to repair. Take care

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)